Ir para o conteúdo principal
Pressure Accuracy Verification for Philips Respironics V200 Esprit
Equipamento Médico
Nem o iFixit nem este guia de reparo são endossados ou afiliados ao fabricante do dispositivo médico.
O que você precisa
  1. Pressure Accuracy Verification for Philips Respironics V200 Esprit, Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 1, imagem 1 %32
    • Put ventilator into diagnostic mode (page 8-21 in Service Manual):
    • Begin with ventilator powered off.
    • Press ALARM RESET and 100% O2 keys located on the right of the front panel key section (as shown in Figure 8-2) for approximately 5 seconds while turning ventilator power on.
    • Touch OK to enter diagnostic mode.
  2. Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 2, imagem 1 %32
    • Connect a patient circuit and analyzer to the ventilator as shown in Figure 8-19 (page 8-21 in Service Manual).
    • Set the analyzer’s function to read cmH2O by setting pressure range to measure at least 120 cmH2O (page 8-21 in Service Manual).
  3. Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 4, imagem 1 %32
    • On the Hardware screen, touch safety (white background, as shown in figure 8-18 on page 8-21 of Service Manual) to energize the safety solenoid.
  4. Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 5, imagem 1 %32
    • On the Hardware screen, touch Exhalation (blue background, seen in Figure 8-18, found on page 8-21 of the Service Manual).
  5. Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 6, imagem 1 %32
    • On the Hardware screen, touch Air (blue background, seen in the top left area of Figure 8-18 on page 8-21 of the Service Manual).
  6. Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 7, imagem 1 %32
    • On the Hardware screen, touch Exhalation and adjust the steps until the analyzer pressure reads 100 +- 5 cmH2O (95 to 105 cmH2O).
    • Information found on page 8-21 of the Service Manual.
  7. Pressure Accuracy Verification for Philips Respironics V200 Esprit: passo 8, imagem 1 %32
    • Check that the Inhalation Pressure and Exhalation Pressure displays on the Hardware screen read within ± 10% of the analyzer display.
    • Example: If the analyzer’s display reads 98.7 cmH2O, 10% of the display is 9.87 cmH2O, and ± 10% of the analyzer display would be 88.83 to 108.57 cmH2O (-10% = 98.7 - 9.87 = 88.83 cmH2O, and + 10% = 98.7 + 9.87 = 108.57 cmH2O).
    • Information found of page 8-21 of the Service Manual.
    • Pressure accuracy test is complete.
Conclusão
If Inhalation/Exhalation pressure readings are out of range at 100 cmH2O, recommended troubleshooting and repair is as follows and can be found in the Service Manual in Chapter 8.7.5 page 8-44:
  1. Check for leaks at circuit connections, test lung, filters, etc.
  2. Check for kinked or cut tubing from inhalation module to SOL4, and from SOL4 to sensor PCB.
  3. Check for kinked or cut tubing from exhalation pressure tap to SOL3, and from SOL3 to sensor PCB.
  4. Check for leaks at the oxygen sensor/oxygen valve connection.
  5. Replace the sensor PCB.
  6. Replace the 3-station solenoid.
Meredith Lee
Membro desde: 05/04/20
169 Reputação
Autoria de 1 guia
Equipe
Cal Poly, Team S1-G4, Zimmerman Spring 2020 Membro de Cal Poly, Team S1-G4, Zimmerman Spring 2020
CPSU-ZIMMERMAN-S20S1G4
Membros da 5
Autoria de 5 guias
0 comentários
Adicione um comentário
Exibir estatísticas:
Últimas 24 horas: 0
Últimos 7 dias: 1
Últimos 30 dias: 4
Duração total: 180
